Controller calibration
« on: December 16, 2020, 06:58:02 pm »
Auto set my calibration..sets it at 66 seems really high..any ideas why it’s so high?

Don Kee

  • Posts: 4878
Re: Controller calibration
« Reply #1 on: December 16, 2020, 07:14:12 pm »
Got a leak free system?
Why don't you have a quick google before making stupid comments?

mac74

  • Posts: 486
Re: Controller calibration
« Reply #2 on: December 16, 2020, 07:40:27 pm »
Ian Shepard is the man to answer this........

Evan

  • Posts: 71
Re: Controller calibration
« Reply #3 on: December 16, 2020, 08:07:18 pm »
Yeh leak free as far as I’m aware...

Spruce

  • Posts: 8462
Re: Controller calibration
« Reply #4 on: December 17, 2020, 08:20:03 am »
Auto set my calibration..sets it at 66 seems really high..any ideas why it’s so high?

If you think its too high then manually reduce it. Set it to 50 and see how it works. If the pump starts to cycle (On/off) then increase it until it stops cycling on the flow rate setting you normally use.

If you are using microbore and the water is cold then the settings need to be higher. We presume you have your flow running connected to your pole hose when you are setting the auto calibration.

If you like a high flow rate then the calibration setting is going to be higher as the pump needs more current to push that volume of water through the hose.
Success is 1% inspiration, 98% perspiration and 2% attention to detail!

The older I get, the better I was ;)

Simon Trapani

  • Posts: 1561
Re: Controller calibration
« Reply #5 on: December 17, 2020, 08:45:51 am »
If he didn’t have the flow running the v11  comes up with ‘err’ for error. Not sure about previous versions.

Edit - Won’t be on earlier versions as I don’t think they had auto calibration.

Ian Sheppard

  • Posts: 1225
Re: Controller calibration
« Reply #6 on: December 17, 2020, 02:26:08 pm »
Auto set my calibration..sets it at 66 seems really high..any ideas why it’s so high?

If you think its too high then manually reduce it. Set it to 50 and see how it works. If the pump starts to cycle (On/off) then increase it until it stops cycling on the flow rate setting you normally use.

If you are using microbore and the water is cold then the settings need to be higher. We presume you have your flow running connected to your pole hose when you are setting the auto calibration.

If you like a high flow rate then the calibration setting is going to be higher as the pump needs more current to push that volume of water through the hose.

Thanks Spruce. That covers it nicely
As in Spruces comments.
Hose ID
Hose Length
Flow rates
Low temperatures

Will all have an impact on the calibration setting  of a control. Each system is a little different and based on preferences such as Flow rates the cal value will be different for each system.

A Cal value of 66 in itself is not and issue particularly if you like to run a higher flow rate. 

Cheers