Sorry to be a pain.i set up a ro with resin in last chamber a few weeks ago.300gpd.out the tap TDS reads 210ppm. Filled and refilled 9 25 litre barrels and then filled up 2 200 litr containers.was reading 002 ppm.not a lot of water has been made has it but now it’s crept up to 009 ppm. I changed the resin in the last chamber but it’s still saying the same.any ideas ?this was a brand new setup that came with my trolley
Let's get some more info.
We have your tap water TDs of 210.
What is the TDs of the pure after the membranes and before it goes to Di?
When processing water how much water goes to waste and how much is pure. In other words we are looking for about 2 liters of waste water to 1 liter of pure.
You have a 300 gpd r/o. Who supplied the system?
How many membranes housings does it have?
You are using the third housing as a Di vessel. For window cleaning you need a bigger one. But you need to sort the problem with the r/o first.